I am very impressed by the NanoV2+4 performance and user interface. However I want to point out the limitation in measuring a low noise amplifier. I would like to measure teh S11 and S21 of an amplifiers with 40 dB gain and output P1dB of +12 dBm; thus an input power of <-28dBm is required. I see the option in Stimulus/CFG Sweep//ADF4350 to reduce the power input to the DUT by 9dB and when I select this and add a 12 dB pad on Port 1 before calibration i measure a pulsed on power of -32 dBm at 1500 MHz. S11 measures correctly at around -20dB *until I connect Port 2 to the amplifier output - it then increases by about 10dB!* This increase in S11 does not occur if I terminate the LNA output. As I connect Port 2 to the amplifier output I see this large change in the indicated S11. This is occurring when the Port 2 input is > +5dBm. My conclusion is that there is a non-linear cross talk from Port 2 to Port 1 when the Port 2 is > +5dBm. I suspect this occurs at the SW/RECV switch at the input of he AD342 mixer. The switch isolation fails at >+5 dBm input from Port 2.
I suggest that Rev 3 of the Nanovna include larger attenuation steps in the source power - perhaps steps of 0, 10, 20, and 30dB. The high level cross talk problem I describe above should be corrected. A specification of maximum power into a Port is needed both from a criteria of damage and of correct operation.