Jim Lux 2025/08/28 06:42
I’m using NanoVNA-Saver and original NanoVNAs, and I have no problem
calibrating at the end of a 30m/100ft transmission line, either with the
NanoVNA by itself, or doing it inside NanoVNA.
I’ve not looked at the code inside the NanoVNA to see how long it allows after
setting the synthesizers to a new frequency, before capturing the next epoch
of samples to measure the values. It’s not “speedy”. I would expect the
synthesizer to settle quite quickly - you’d have to look at the schematic and
code to see what the loop filter time constant is (the loop filter is
programmable, and it depends, also, on the reference frequency/step size of
the PLL)
As I recall, it collects a millisecond worth of samples (maybe faster on the
VNA2), which sets the basic 1 kHz measurement bandwidth.
The delay through a 30m transmission line, one way, is on the order of 150 ns
one way.
So the two way delay is 300 ns. Let’s assume that the audio ADC starts
sampling instantly after the frequency change is done - at 96 ksps, that’s
about 10 microseconds. So out of the hundred samples taken to form the I/Q
value, you might have just the first one corrupted, maybe.
I don’t think that’s your problem.
> On Aug 28, 2025, at 00:40, Glenn n6gn via groups.io
<n6gn=sonic.net@groups.io> wrote:
>
>
>
>
> I'm trying to set a measurement/calibration plane out at the end of two 15m
(50 foot) test cables. While a normal nanovnasaver measurement with
calibration at the SMAs is OK, a calibration made with the SW out at the ends
of these cables is not. I think there is insufficient dwell time after
frequency point has changed to allow a stable reading in this high group
delay environment. Measurement BW seems to be down at 1 kHz so perhaps this
is influencing the required settling time.
>
>
>
> On the VNA itself Bandwidth is indicated as 6.2 kHz 400 P/s with not
adjustment possible. In the SW there seems to be no adjustment for this
calibration situation which seems not to be appropriate.
>
>
>
> I can make a good calibration if I do not use nanovnasaver but only the on-
board VNA calibration tool. vnasaver seems to not provide enough dwell to
allow a new measurement condition to stabilize on these longer cables.
>
>
>
> What is the correct way to slow each frequency point measurement setting,
prior to the calibration measurement, in order to achieve good calibration
with vnasaver?
>
>
>
> Glenn n6gn
_._,_._,_
* * *