CT2FZI 2020/06/13 11:19
What about using nanovna-saver software, calibrate using big start and stop
frequencies, but with extra segmentation (101 points X N segments) to give
it extra detail?
Then, after loading calibration, one can sweep only in the frequencies of
interest, and even reduce the segments to have faster scanning.
The calibration needs to be done once, and for every cable/adaptor
combination, but you would only do it once... Then it's a matter of
recalling the correct calibration and measure.
Suggestion:
50Hz-53MHz (30 segments)
3Mhz-30Mhz (30 segments)
49MHz-71MHz (10 segments)
135-150MHz (20 segments)
420-450MHz (20 segments)
...
This will work ok and you can skip calibration #2 if you want.
Then load calibration, and just sweep for part or all of the calibration
start and stop frequencies...
For antenna analysis I am pretty sure that one would only need one
calibration, ie, without minding the adapters, or cables...
*73 de Luís, CT2FZI*
<http://goog_717545661>
On Sat, 13 Jun 2020 at 10:52, ok1vaw <ok1vaw@volny.cz> wrote:
> I suppose too, that there could be an interpolation model found for
> approximating two port calibration matrix after SOL one port calibration
> find or maybe for the SOLT too. The need for recalibration on every sweep
> change is bad. Another way is to make a time-consuming but one and only
> full band calibration on the beginning of the setup and then use only
> frequencies, where the calibration was done.
> From my point of view, the sweep setting with "segment" I would like to
> see changed to number of steps or frequency step.
>
>
>
>